the mode of connection dentalimplant and abutment fall under product structure andcomponents?

您还未登录!暂时最多只可查看 1 条回答

登录! 还没有账号?去注册

si4520 一阶会员 用户来自于: 广东省深圳市
2023-09-18 20:06

Yes. The mode of connection between dental implant andabutment is among the contents that should be specified in theproduct structure and components. The modes of implant-abutmentconnection is mainly classified into external and internal connectiontypes, and can also be classified into internal square connection.external hex connection,external octagon connection,internalhexagon connection and Morse taper connection by geometricshapes. Based on different modes of connection, the administrative counterpart should respectively carry out the relevant studies on theverification of system compatibility such as the taper fit of internalconnection of the implant, the fit clearance between the implant andthe abutment, the torsional property, the tightening torque and thefatigue limits.